fluence sysbacip:It is fresh Raw meal not mixed with recirculation dust.
Hello Sheikh,
In that case, the influence of the recirculation dust is irrelevant because ultimately the input is fresh raw meal (LSF~91) and the output is clinker (LSF~93).
As Ovancantfort has already mentioned the observed difference in LSF is most likely due to sampling and/or analysis errors in either the raw meal or the clinker, or both.
You don't need to have very large analysis errors to result in a difference in LSF of only 2, if the biases happen to be in the right directions.
For example;-
If a sample of raw meal has an actual LSF of 92 and the following analytical biases;-
SiO2 +0.10
Al2O3 +0.05
Fe2O3 +0.05
CaO -0.10
the measured LSF will be 91.
And if a sample of clinker has an actual LSF of 92 and the following analytical biases;-
SiO2 -0.10
Al2O3 -0.05
Fe2O3 -0.05
CaO +0.10
the measured LSF will be 93.
So you can see that even relatively small analysis errors can explain the observed difference of roughly 2 LSF units between the measured raw meal and clinker. And this is just considering the analytical errors. Analytical errors are usually small compared to sampling errors, so a combination of both errors would easily explain the observed difference in LSF.
Regards,
Ted.